When a bare thermocouple is in contact with a metal surface, a seebeck voltage is developed between the thermocouple and the metal.stefanbanev wrote:Dear sparkyy0007,
I did not follow the technical issues/details you've pointed out; besides, I have no sufficient expertise with thermocouples. May you summarize the possible effects of these issues; how it could effect the Energy out/in ratio through all run, may you assess the ranges as non-favorable to e-cat and also favorable as well.
Thanks
The hotter the metal, the higher that voltage will be. This is not a problem with a single thermocouple because that voltage is not seen by the meter.
If however 2 thermocouples are connected to the same instrument and both thermocouples are at different temperatures on the same metal substrate, a differential voltage will develop between the thermocouple inputs and this causes problems with offset and intermittent contact errors.
The solution is simply to use floating/isolated thermocouples, cheap and easy solution, standard industry practice.
This may not be a problem in this demonstration, but it is a possibility.
I won't comment on it's implications of the "excess energy " generation however, without knowing accurate temperatures, how can calculate anything.